Pelton & Crane
Model of Sterilizer:
Delta 10
Problem Title:
Status of lights
Problem Description:
The status of the lights. The overheat and heat light never energize. The fil light does but you can hear the solenoid energizing and deenergizeing numerous times before it changes to sterilize mode.

Solutions

1. It is normal that the water fill solenoid on/off/on/off during the water filling period.

However, the heating element status light should be on in the heating period, and on/off/on/off in the filling period.

Did you check the heating elements? Disconnect the main power, check the Ohm reading of the heaters. Howerver, probably the PCB board is out of work.