MDT Sterilizer
Model of Sterilizer:
Harvey Hydroclave SC8
Problem Title:
not pressurizing enough did not pass Spore Test
Problem Description:
This question is about Chemiclave 5000. need assistance on what to do and what to repair and replace. Please..
Reasons
1. Check for door gasket leakage and for any air bubbles rising from metering valve while running a cycle.

Solutions

1. Adjust door to help prevent leakage from door gasket. Replace door gasket and or metering valve if leakage exist after adjusting door.

The technical information is just to help sterilizer users as reference to do their trouble shooting.