Sterilizer Manufacturer:
Market Forge
Model of Sterilizer:
Market Forge STM-E
Problem Title:
not reaching temperature
Problem Description:
Our pressure is at 15-16 psi but our temperature is only getting to 120 degrees? What should we do?
Reasons
1. Temp gage may be defective.

Solutions

1. You would need to use a max register thermometer to confirm the temp reading in the chamber. I would guess that your pressure gage is reading correctly and more than likely, the temp gage will need to be replaced.