Problem Title:
Statim 2000
Problem Description:
The exhaust hose keeps coming off the waste bottle when we start the Statim machine.
The diagram in the operator's manual is not clear.
Can you help?
Reasons
1. The nut needs to be tightened or is stripped,the ferral is missing from the nut assy.

Solutions

1. If the fitting cannot be tightened or the ferral that crimps the tubing is missing or damaged, replace the fitting and before installing, cut 1 inch off the end of the tubing. Make sure the tubing is pushed all the way into the fitting, hand tighten and then using a pair of pliers, tighten another 1/2 to 3/4 turn.

The technical information is just to help sterilizer users as reference to do their trouble shooting.