Sterilizer Manufacturer: Midmark |
Model of Sterilizer: M7 |
Problem Title:
Midmark Ritter M7 sterilizer Fill/Vent function does not work properly |
Problem Description:
There is an issue wit the fill. It is very slow, and I am not sure what I am supposed to do. |
Reasons |
1. 1) Fill/Vent valve is clogged with foreign matter 2) Fill/Vent valve is not adjusted properly 3) Condensing tank is low on distilled or demineralized water 4) Tubing is clogged |
Solutions |
1. Take out the fill vent valve and clean it real well. |
2. Buy a new one. |
3. Add water |
4. Take out the tubbing from the reservoir, and clean real wel.. |
|
|||||