Statim 2000
Problem Title:
SCICAN Statim error # 2
Problem Description:
I've had the Statim 2000 for a while, and all of a sudden I started to get error 2. Why?
Reasons
1. SCICAN Statim error # 2

The casseette temperature failed to increase from 95C to 100C within a time-out period

1) Unit will normally have a major steam leak from the cassette, replace the seal, lid or tray.
2) If the cassette is not leaking, unit is reading incorrect chamber temperature, a calibration cassette is required to check chamber temperature. Shop repair needed.
3) An extremely large load (Statim 5000).

The technical information is just to help sterilizer users as reference to do their trouble shooting.