SCICAN Statim error code: 3
Problem Description:
SCICAN Statim error code: 3
Reasons
1. SCICAN Statim error code: 3

The cassette has failed to pressurized and achieve a temperature of 110C within a time-out period

1) Check steam leaking, replace seal, lid or tray
2) If no leak are visible, check solenoid valve and plunger sticking in the plunger tube
3) Check for the constant power to the solenoid J1-7 and J1-8 should be zero volts with cassette out of the unit.
4) Check the check valve (one way) and pressure relief valves are not leaking

The technical information is just to help sterilizer users as reference to do their trouble shooting.