Pelton & Crane
Model of Sterilizer:
Problem Title:
DELTA 8/10 Sterilizer H2O - Unit will not fill properly
Problem Description:
DELTA 8/10 Sterilizer H2O - Unit will not fill properly
Reasons
1. DELTA 8 or 10 Sterilizer H2O - Unit will not fill properly

The H2O alarm is a direct response from the Surface Sensor. The Surface Sensor is a temperature sensitive device that is fastened to the underside of the chamber. It monitors the temperature of the chamber and can distinguish, by the increase in temperature, when the water level has receeded beyond a safe level. At this point (159C), the Surface Sensor opens and H2O displayed. This is to prevent a boil dry situation that can ruin the heating element and chamber. This is caused by one of the following conditions:

A) The chamber will not fill properly
B) The unit is not properly leveled
C) The chamber leaks under pressure
D) Defective or loose Surface Sensor or defective cable.
2. 1) Close the door and start an Unwrapped cycle, when the FILL message disappeared, open the door to see the water level should be half way up to the dam?

If yes, see H2O problem caused by reason of "Chamber loses excessive water"

If no, then you need to check the Filter inside chamber and reservoir