DELTA 8 10 sterilizer Error S1
Problem Description:
DELTA 8 10 sterilizer Error S1
Reasons
1. DELTA 8 10 sterilizer Error S1

Fill/Vent solenoid coil or circuit open
2. 1) Check the Fill Vent solenoid Ohm, is it 16 ohm?
2) check the connection of this solenoid on the Power PCB board
3) Check the ribbon cable