Model of Sterilizer:
Tuttnauer 2540EHS
Problem Title:
Tuttnauer Heat Up Problem
Problem Description:
Low Heat Message is displayed
Reasons
1.

Low Heat

-- Refers to the temperature in the Chamber before sterilization has begun.

This message is displayed if the programmed sterilization temperature is not reached within 50 minutes of the start of the instrument cycle or 80 minutes for the liquid cycle.

Solutions

1. Clogged Air Jet
Very low voltage delaying heat up
Bad Heating Elements or heater connection
Safety Thermostat opening prematurely, turning off Heating Elements - only on units with control proms earlier than T93N3
Air Outlet Valve stuck close