Sterilizer.com Sterilizer Parts Technical Home FAQ Search Submit a Question Your Local Service Login
Enter question keywords
to search

Sterilizer Manufacturer:
Tuttnauer
Model of Sterilizer:
Problem Title:
Tuttnauer circuit breaker or fuse problem
Problem Description:
Circuit breaker or fuse trips each time unit is turned on
Reasons
1. If the problem of either Power Supply fuse or Dry Dump fuse, please check relative issue first

Solutions

1. Check that the voltage is in the correct range for that unit.

The proper voltage range for 110 volt machines is between 110 and 125. The proper range for 220 volt machines is between 220 and 235 volts.

Check for a ground short at the Heating Elements

Check for damaged or shorted wires throughout the unit

The technical information is just to help sterilizer users as reference to do their trouble shooting.
no guarranty implied. For each specific technical help to do with sterilizers
or autoclaves, go to www.sterilizers.com ... or call 1-800-801-9934 Contact Us