Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
Air Outlet Valve Problem Tuttnauer E/EA/EK/EAK |
Problem Description:
Tuttnauer air outlet stays colsed all the time |
Reasons |
1. The solenoid does not work either caused by itself or the control signal |
Solutions |
1. Check between TP14 and TP1 for a 0 to 1 volt DC reading. If TP14 never shows between 0 and 1 volt this would indicate a control problem Check for a broken orange or black wire going to the Solenoid Coil Remove the Solenoid Coil and check for a short in the Solenoid Coil Remove the connection box from the Solenoid Coil and check that the connectors are in place and making good contact with the terminals on the Solenoid Coil. Check that the Plunger Assembly is clean and not sticking Check that the Plunger Housing is not deformed causing the plunger to stick |
|
|||||