Model of Sterilizer:
Problem Title:
Air Outlet Valve Problem Tuttnauer E/EA/EK/EAK
Problem Description:
Tuttnauer air outlet stays colsed all the time
Reasons
1. The solenoid does not work either caused by itself or the control signal

Solutions

1. Check between TP14 and TP1 for a 0 to 1 volt DC reading. If TP14 never shows between 0 and 1 volt this would indicate a control problem

Check for a broken orange or black wire going to the Solenoid Coil

Remove the Solenoid Coil and check for a short in the Solenoid Coil

Remove the connection box from the Solenoid Coil and check that the connectors are in place and making good contact with the terminals on the Solenoid Coil.

Check that the Plunger Assembly is clean and not sticking

Check that the Plunger Housing is not deformed causing the plunger to stick