Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
Cut-Out Theremostat Problems of Tuttnauer E/EA/EK/EAK |
Problem Description:
Red reset button will not reset |
Reasons |
1. Either the Cut-Off is always activated or it is damaged |
Solutions |
1. Make sure the Chamber has cooled. If the Chamber remains hot the Cut-Out Thermostat can not be reset. |
2. Use a sharp pointed object, like the point of a pencil or pen to fully depress the reset button if necessary. When reset a small click can be detected, the button however will not stay in. |
3. If power is not restored to the autoclave then unplug the unit and remove the two power wires going to the Cut-Out Thermostat. Use an ohm meter to check continuity across the terminals |
4. If continuity can not be restored across the terminals of the Cut-Out Thermostat by pushing in the red reset button then replace the Cut-Out Thermostat |
|
|||||