Scican Statim 2000 sterilizer noise loudly
Problem Description:
Scican Statim 2000 sterilizer noise loudly
Reasons
1. The Statim 2000 has a big noise coming from the Dry Cycle.
2. People will be easy to think this problem is from the Air Dry Pump. However mostly this problem is caused by the one way Valve on the boiler. If this valve is out of work, it will be 2 status.
1) Always open. In this case, you will see the Dry Air Pump is damaged too with water in.
2) Always closed/almost closed. In this case, the valve will block the Dry Air flow to the cassette.
The easy way to distinguish which case happened, is in the second case (closed), the machine will still work in most cycle except the loud noise during the Air Dry.

The technical information is just to help sterilizer users as reference to do their trouble shooting.