A6- 5560 Magnaclave
Problem Description:
A6- 5560 Magnaclave
Reasons
1. Maintenance and Trouble shooting
2. Sorry the correct series number is A6-5520. The machine was not able to build pressure. the temperature and pressure gauges could not go to 134 C Degree and 30 PSIA
3. The problem was caused by a leak in the bellow house. After fixed the house problem, the machine works well.
Magnaclave machine has a designing problem in the exhausting pipe. During the venting time, it is very easy for the hot steam water flow out of the reservoir and some water will flow inside the machine. No a very good way to fix this problem. Keep the filter on the end side of the cooling pipe as clean and possible. and try to make the filter in the center position on the end side of reservior.