Tuttnauer EHS "Low temp" is displayed
Problem Description:
Tuttnauer EHS "Low temp" is displayed
Reasons
1. In few cases, the Tuttnauer EHS sometimes failed with Cover, but works OK without Cover
2. Tuttnauer EHS "Low temp" is displayed

low temperature and low pressure

maybe there is a leakage from the chamber or jacket: Fix the leakage
maybe the faulty of heating element: replace the heating element
maybe the pressure switch: replace the pressure switch
maybe the cut-off thermostat disconnect the heating element: reset the cut-off or replace it.

On 3870 machine: maybe the contactor faulty: replace the contact

lower water level electrode does not sense water

Contaminated electrode: replace the electrode
Faulty electrode: replace the electrode
Low or no water inlet and water pump operates too often:
water inlet is clogged: clean the piping
faulty water pump: replace water pump
water inlet non-return (one way) valve is stuck: clean or replace the valve