Sterilizer Manufacturer:
Pelton & Crane
Model of Sterilizer:
Delta 10
Problem Title:
Pelton Crane Delta 10 Pressure error code
Problem Description:
I get a pressure error code.

I have changed both the sensor and the board as was suggested, now the pressure indicator no longer fluctuates, but now the pressures goes to 240+ and shuts down and still gives the pres error code!!! What do I do next?
Reasons
1. What is the error code now?

Solutions

1. What is the pressure error code now. P1, or P2. It sounds that your machine pressure goes over 241KPA (I guess the pressure unit is KPA, impossible PSI) before its temperature reading 134C, a calibration needed on the board. Where did you get the new board?