Sterilizer Manufacturer:
Scican Statim
Model of Sterilizer:
Statim 2000
Problem Title:
Statim 2000 Error 26
Problem Description:
how do you calibrate the boiler.

Solutions

1. 1) Turn OFF main power and remove its cover
2) Short together PINs W1 on the main PCB board.
3) Turn ON the power.
4) the LCD should display CALIBRATION with:
2 digits in the upper left, 2 digits in the upper right.
the lower left will show the temperature difference between boiler and chamber. the lower rightside shows the chamber temperature.
5) insert cassette
6) HOLD down the UNWRAPPED and press START
7) the 2 digits in the upper left change to 00> and a message CALIBRATING* shows up
8) it takes about 5 minutes, a long beep will sound. The upper left will change to 2 digits again.
9) press STOP to allow the Statim to vent
10) when it shows PRESS STOP TO RESET, the press STOP
11) it will display SELECT A CYCLE.
12) turn OFF the power, and remove the JUMPER on PINs W1.
13) put the cover back.
14) turn ON the power,
all set, you have the machine calibrated.

The technical information is just to help sterilizer users as reference to do their trouble shooting.