Problem Title:
Scican Statim 2000 -error 16
Problem Description:
We get error 16 and the cycle will not go above 97 degrees. then it stops. I am in the middle of a clinic trying to do 35 surgeries a day. Please help.
Reasons
1. Either the water pump or pump filter out of work, or
the boiler out of calibration

Solutions

1. Check the water pump and its filter first. If both of them are working well, then you need to recalibration the boiler.

The problem is its temperature reading is wrong
2. FYI

visit Statim 2000 calibration

The technical information is just to help sterilizer users as reference to do their trouble shooting.