Sterilizer Manufacturer:
Scican Statim
Model of Sterilizer:
Statim 2000
Problem Title:
Statim 2000 sterilizer cut-off temperature
Problem Description:
what is the cut-off temperature in celsius or fahrenheit for the thermal fuse?

Solutions

1. No specific data for cut-off temperature in celsius or fahrenheit for the thermal fuse.

However, if the boiler temperature over 170C (about 338F), or the chamber temperature over 147C (about 297F), or the thermofuse broken, then you will get error message Service Needed.

The technical information is just to help sterilizer users as reference to do their trouble shooting.