Tuttnauer EHS vacuum valve (52) always open
Problem Description:
Tuttnauer EHS vacuum valve (52) always open
Reasons
1. Tuttnauer EHS vacuum valve (52) always open (3870EHS has 2 valves)

Check the valve ground short connection at JP10-4
Check the DIG-T2 board TP8 voltage TP8<1V the valve will open
Check the solenoid on the valve, if the voltage of coil is over 10V, then the ANL-T2 board is faulty

The technical information is just to help sterilizer users as reference to do their trouble shooting.