Sterilizer.com Sterilizer Parts Technical Home FAQ Search Submit a Question Your Local Service Login
Enter question keywords
to search


Sterilizer Manufacturer:
Midmark
Model of Sterilizer:
M7
Problem Title:
M7 Speedclave not working properly
Problem Description:
Hi this is J.B.the Piercer I bought a refurbished M7 from u in Nov. 2006, the unit is not coming up to temp 265 degrees for 15 min. the over flow is steeming not alowing unit to work properly.I think the Barnes&Jones Part type D 0-80 PSIG 6baror my research part # D222 is faulty. Please contact me and let me know what to do?
Thankyou J.B.

Solutions

1. You should try by replacing the bellow housing. Follow the instructions here
http://www.autoclave-parts.com/Z-midmark/M7/M7-Door-Bellows-Page.html
2. You should try by replacing the bellow housing. Follow the instructions here
http://www.autoclave-parts.com/Z-midmark/M7/M7-Door-Bellows-Page.html
3. Or this bellows

The technical information is just to help sterilizer users as reference to do their trouble shooting.
no guarranty implied. For each specific technical help to do with sterilizers
or autoclaves, go to www.sterilizers.com ... or call 1-800-801-9934 Contact Us