DELTA 8
Problem Description:
DELTA 8
Reasons
1. Machine shows FAIL after 10-20 minutes
2. This is a very new machine. But never success in any cycle. After running 10-20 minutes, it shows FAIL on the screen.

RULE, if it is too long time to get the temperature of 110 C during the heating period, it will show this error message.

Reason: the calibration is out of work. The real temperature inside the chamber is over 110C, which the digital reading is still below 110C. It can be fixed by calibration.

Pelton & Crane does not have instruction to show customer how to calibrate its machine, you need to change the PCB board complete.

The technical information is just to help sterilizer users as reference to do their trouble shooting.