Tuttnauer EHS float switch in the reservoir is always ON or OFF
Problem Description:
Tuttnauer EHS float switch in the reservoir is always ON or OFF
Reasons
1. Tuttnauer EHS float switch in the reservoir is always ON or OFF

1) make sure it can move free inside the reservoir
2) check all the wire connection to the float
3) check the ohm reading of the float in On/Off position to make sure getting different ohm reading
4) If the above is OK, then either ANL-T2 or DIG-T2 board is out of work

The technical information is just to help sterilizer users as reference to do their trouble shooting.