Tuttnauer EHS electrode does not sense water
Problem Description:
Tuttnauer EHS electrode does not sense water
Reasons
1. Tuttnauer EHS electrode does not sense water enve there is enough water in the jacket.

1) do the In-Out test, to make sure the connection values: Open should be in the 255 side, short should be in the 0 side
2) clean the electrode
3) check the wire connection to the ANL-T2 board on JP11

The technical information is just to help sterilizer users as reference to do their trouble shooting.