Sterilizer Manufacturer:
Market Forge
Model of Sterilizer:
Market Forge STM-E
Problem Title:
Loses pressure toward end of run, water is present in the chamber after run.
Problem Description:
Loses pressure about 2-4 minutes before the end of a 15 minute run, water is present in the chamber after run. Replaced the low-water switch and does not resolve the problem.
Reasons
1. Losing pressure at the end of a run is usually caused by not enough water in the chamber to begin the cycle, leakage from steam trap, drain,door gasket,safety or possibly the fast exhaust solenoid.

Solutions

1. Check the reasons and inspect for any of those problems.

The technical information is just to help sterilizer users as reference to do their trouble shooting.
no guarranty implied. For each specific technical help to do with sterilizers
or autoclaves, go to www.sterilizers.com ... or call 1-800-801-9934 Contact Us