Sterilizer Manufacturer:
Tuttnauer
Model of Sterilizer:
Problem Title:
Tuttnauer M insufficient water fill
Problem Description:
Tuttnauer M insufficient water fill
Reasons
1. Tuttnauer M insufficient water fill

1) No water in the reservoir: fill the reservoir
2) Drain valve open: close
3) MPV (Multi Purpose Valve) not in fill position
4) Wrong know the the MPV
5) MPV jammed
6) MPV or tubing clogged

The technical information is just to help sterilizer users as reference to do their trouble shooting.