Sterilizer.com Sterilizer Parts Technical Home FAQ Search Submit a Question Your Local Service Login
Enter question keywords
to search

Sterilizer Manufacturer:
Pelton & Crane
Model of Sterilizer:
OCR
Problem Title:
Heating on OCR autoclave
Problem Description:
I am working on this unit and it is getting heat but after 1 1/2 hours the temp is only at 240 and the psi is about 8. I don't know if the heating elements need to be replaced or if I need to look at something else.

If it is the heating elements what is the part # and does it include all 3 elements or 1 each.
Reasons
1. Is this an original machine or does it have an autovent function?

Solutions

1. You can test the heater elements ohm readings. They should read about 11-12 ohms

The technical information is just to help sterilizer users as reference to do their trouble shooting.
no guarranty implied. For each specific technical help to do with sterilizers
or autoclaves, go to www.sterilizers.com ... or call 1-800-801-9934 Contact Us