Prestige 2100 Vacuum Error Code E03
Problem Description:
Prestige 2100 Vacuum Error Code E03
Reasons
1. Prestige 2100 Vacuum Error Code E03

Air bleed was not successful

Autoclave has failed to obtain a balance between boiler and chamber PT100 probes within the time limit

1) Very low applied voltage (less than 188V for 230V)
2) Gross steam leak
3) Boiler not filling with water
4) Air bleed tubes blocked
5) Air bleed solenoid valve failed
6) Air bleed solenoid blocked
7) Check chamber probe is correctly inserted
8) Check that both PT100 probes have a similar resistance at ambient temperature
9) Check power control module