Sterilizer.com Sterilizer Parts Technical Home FAQ Search Submit a Question Your Local Service Login
Enter question keywords
to search

Sterilizer Manufacturer:
Tuttnauer
Model of Sterilizer:
All of 'E' models
Problem Title:
Tuttnauer 2340E models Pressure Problem
Problem Description:
High Pressure message is displayed
Reasons
1.

High Pressure

-- Refers to Chamber pressure after sterilization temperature has been reached. If the Chamber pressure rises 10psi above the pressure required for sterilization the cycle will abort.

Solutions

1. 1. Sterilization temperature has been set above 273
2. Heating Elements are not centered or tightened properly.
3. Heating Elements remaining on instead of cycling on and off.
4. Pressure or Temperature calibration is incorrect Turn off the power and disconnect and then reconnect all cables and connectors inside and outside the Electronic Box making sure that there are no bent or broken pins and all connections are solid.

Verify that the pressure and temperature reading are correct by using a mechanical pressure gauge and an independent thermometer.
If the displayed pressure is not accurate then recalibrate the Pressure Sensor

If the displayed temperature is not accurate then recalibrate the Temperature Sensor

The technical information is just to help sterilizer users as reference to do their trouble shooting.
no guarranty implied. For each specific technical help to do with sterilizers
or autoclaves, go to www.sterilizers.com ... or call 1-800-801-9934 Contact Us