Tuttnauer E/EA/EK/EKA Safety Thermostat Problems
Problem Description:
Tuttnauer There is no continuity across the safety thermostat
Reasons
1. see solutions for detail

Solutions

1. To correctly check continuity unplug the unit and remove the two power wires from the Safety Thermostat, use an ohmmeter to check across the terminals for continuity

Make sure the Chamber has cooled. If the Chamber remains hot the Safety Thermostat will not automatically reset.

If the Safety Thermostat will not automatically reset and restore continuity across it’s terminals then replace the Safety Thermostat