Problem Title:
Tuttnauer Circuit breaker or fuse problem
Problem Description:
there is no output from the circuit breaker or fuse
Reasons
1. Check the fuse problem first, if there is a fuse problem with either Power Supply or Dry Pump

Solutions

1. Check for correct voltage at wall outlet

The proper voltage range for 110 volt machines is between 110 and 125. The proper range for 220 volt machines is between 220 and 235 volts.

Check for bad Line Cord

Check for a faulty power receptacle at the back of the autoclave

Check for voltage on the input side of the Circuit Breaker or Fuse

If this is a fused unit then, check that the Fuse is not blown A visual inspection may not show a blown Fuse, it is always recommended that the Fuse be checked with an ohmmeter.

Check the Fuse Holder for continuity both in and out of the holder.

Replace the Fuse Holder if necessary.

If this unit has a Circuit Breaker then check that the Circuit Breaker is in the on position

If the unit has a Circuit Breaker and all the above are okay then replace the Circuit Breaker

The technical information is just to help sterilizer users as reference to do their trouble shooting.