Problem Title:
Tuttnauer circuit breaker or fuse problem
Problem Description:
Circuit breaker or fuse trips while unit is running
Reasons
1. If the main problem is Power Supply or Dry Pump fuse, please check them first

Solutions

1. Check that the voltage is in the correct range for that unit.

The proper voltage range for 110 volt machines is between 110 and 25. The proper range for 220 volt machines is between 220 and 235 volts.

Check the resistance of the Heating Elements, of particular interest would be a reading to ground. Check when the unit is hot but not plugged in

Check for damaged or shorted wires throughout the unit

The technical information is just to help sterilizer users as reference to do their trouble shooting.