Air Outlet Valve Problem Tuttnauer E/EA/EK/EAK
Problem Description:
Air Outlet blows off during sterile cycle
Reasons
1. some settings or readings are wrong

Solutions

1. Check that the temperature setting is not set any higher then 273F (134C)

Check that the Heating Elements are turning on and off properly

Check that the Heating Elements are installed properly
2. Recalibrate the temperature sensor
3. Recalibrate the temperature sensor (pressure transducer)