Tuttnauer 2340 E/EK/EA/EKA
Problem Title:
Tuttnauer - 2540E Air Outlet Valve Problem
Problem Description:
Tuttnauer air outlet stays open all the time
Reasons
1. either the solenoid control signal wrong or the solenoid itself out of order

Solutions

1. Check between TP14 and TP1 the reading should be between 3.5 and 5
volts DC when the valve is off and 0 to 1 volt when the valve is on.

If TP14 is showing 0 to 1 volt constantly this indicates a control problem

If TP14 is between 3.5 and 5 volts and the valve is open then check for a ground short in the black wire going from the solenoid to the
Electronic Box

Remove the Solenoid Coil and check across the terminals for a short Take apart the Plunger Assembly and clean any debris that may be stopping the Plunger from sliding back and forth.

Check the housing of the Plunger for any nicks or distortions that can cause the Plunger to become stuck.

The technical information is just to help sterilizer users as reference to do their trouble shooting.