Pelton & Crane OCR- Can't fill the chamber with water.
Problem Description:
* Remove the trays out of the chamber.
* Remove the rest tray from the chamber.
* Remove the filter form the back of the chamber (the filter looks like an ice cream cone, about 2" long, and its color is either gold or black). The way to remove the filter is wiggle it from left to right, while pulling it towards your belly.
* try to fill now.
* if it's working, replace the filter.
Reasons
1. Filter is cloged up or dirty.

Solutions

1. You need to repalce the filter

The technical information is just to help sterilizer users as reference to do their trouble shooting.