Problem Title:
Statim 2000 Error 19
Problem Description:
Fault #19, the steam generator calibrationis invalid. A new calibration is required. How do I do this.
Reasons
1. A calibration needed

Solutions

1. 1) Turn off the main power
2) Remove the top cover
3) Short together the Pins W1
4) Turn on the power
5) LCD shows CALIBRATING
Two digits on Upper Left, two digits on the Upper Right
The Lower Left shows the temperature difference between boiler and chamber
The Lower Right shows the chamber temperature
6) Insert cassette
7) Hold down the UNWRAPPED button and press START
8) an asterisk * will show after CALIBRATING
9) Takes about 5 minutes, a long beep will sound
10) Press STOP to vent
11) After LED shows PRESS STOP TO RESET, press STOP again
12) You get a new message SELECT A CYCLE
13) All set

You may get a copy of technical manual for reference.