High press is displayed
Problem Description:
High press is displayed
Reasons
1. "High press" is displayed

Pressure is high and the temperature is low
Slow exhaust or drain lines are clogged, such as:
Clogged steam trap: replace the steam trap
Needle Valve closed: close it completely then open half a turn
Water outlet pipe clogged: clean or replace the pipes
Slow exhaust valve is stuck: clean or replace the valve (74)
2. This message is only applied to Tuttnauer EHS

The technical information is just to help sterilizer users as reference to do their trouble shooting.