Low press is displayed |
Problem Description:
Low press is displayed |
Reasons |
1. Low pressure and High temperature Exhaust or slow exhaust outlets are leaking: fix the leakage Temperature or pressure out of calibration: recalibrate them Air inlet valve (43) is stuck open: clean or replace the valve Vacuum valve (52) is stuck open: clean or replace the valve Fast exhaust valve (73) is stuck open: clean or replace the valve Slow exhaust valve (74) is stuck open: clean or replace the valve Steam valve (93) is stuck open: clean or replace the valve. |
|
|||||