Problem Title:
Low press is displayed
Problem Description:
Low press is displayed
Reasons
1. Low pressure and High temperature

Exhaust or slow exhaust outlets are leaking: fix the leakage
Temperature or pressure out of calibration: recalibrate them
Air inlet valve (43) is stuck open: clean or replace the valve
Vacuum valve (52) is stuck open: clean or replace the valve
Fast exhaust valve (73) is stuck open: clean or replace the valve
Slow exhaust valve (74) is stuck open: clean or replace the valve
Steam valve (93) is stuck open: clean or replace the valve.