Model of Sterilizer:
Problem Title:
High Temp is displayed
Problem Description:
High Temp is displayed
Reasons
1. "High Temp" is displayed

High temperature and low pressure
Steam trap is leaking: fix the leakage
Needle valve is leaking: fix the leakage
Temperature sensor PT100 is faulty: replace the sensor
Pressure sensor is faulty: replace the sensor
Temperature or pressure reading out of calibration: recalibrate
Exhaust lines, chamber or jacket are leaking: fix the leakage
Faulty steam trap: replace the trap
needle valve open: close it completely then open half a turn