| Sterilizer Manufacturer: Tuttnauer |
| Model of Sterilizer: |
|
Problem Title:
EHS "Low heat" is displayed |
|
Problem Description:
EHS "Low heat" is displayed |
| Reasons |
| 1. EHS "Low heat" is displayed Piping leaking: fix the leakage Faulty heating elements: replace the heating elements Faulty pressure switch: replace the pressure switch Faulty cut-off thermostat (3870EHS case): repalce the thermostat Lower water electrode does not sense water: Faulty electrode: replace the electrode Faulty water pump: repace the pump Water inlets is clogged: clean the pipes Air inlet valve (43) is stuck open: clean or replace the valve Vacuum valve (52) is stuck open: clean or replace the valve Fast exhaust valve (73) is stuck open: clean or replace the valve Slow exhaust valve (74) is stuck open: clean or replace the valve Steam valve (93) is stuck open: clean or replace the valve |
|
|
|||||
|
The technical information is just to help sterilizer users as reference to do their trouble shooting. no guarranty implied. For each specific technical help to do with sterilizers or autoclaves, go to www.sterilizers.com ... or call 1-800-801-9934 Contact Us |