Problem Title:
midmark model 9 intermittent error code 5 i have removed the bellows and it is clear any help would be appreciated
Problem Description:
i keep receiving error code 5 intermittently and the bellows is clear
Reasons
1. The machine shuts down during sterilization cycle
2. sterilizer overloaded
3. bellows malfunctioning (closing early)
4. sterilizer potentionmeter out of calibration and/or dirty temperature sensor
5. temperature sensor malfunctioning

Solutions

1. Check that pressure vessel is not overloaded with heavy linen packs.
instruct operator to reduce load size
2. Bellows should close at approx 212F (100C). Replace bellows
3. Clean temperature sensor, then check calibration of using pressure gauge test harness.

Perform temperature/pressure potentionmeter adjustment.
4. check calibration

replace temperature sensor assembly and recalibrate.

The technical information is just to help sterilizer users as reference to do their trouble shooting.