Problem Title:
Tuttnauer 2540 EKA Time on display cant be changed
Problem Description:
When Tuttnauer is started printer begins to run. On display you see after the Version the line 134 05 00 .
OO is underlined, you cannot change anything with all switches. Is there a possibility to reset. Pressing STOP when switching on does not reset anything.

Solutions

1. At beginning the 134 05 00 means as follows:
134 is the temperature in F
05 is the time in minute
00 is the chamber pressure at the beginning.

What do you want to change? The customization keys are different.

Keep pressuring the STOP button until you can see an enter key like 102 (the last digital may be different), then use the UP/DOWN arrow to make it as 105, the press the STOP button again. Now you are in the customizatin mode, keep press the STOP button until the parameter, which you want to change show up, then use the UP/DOWN button to set as what as you want.

If you want do reconfigure some more technical settings, please call me for help. It is not recommended for end user to do some technical settings, such as calibration. My phone number is 516-489-3855 asking for Tuttnauer Assistance.


The technical information is just to help sterilizer users as reference to do their trouble shooting.