Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
Tuttnauer EHS is not drying properly |
Problem Description:
Tuttnauer EHS is not drying properly |
Reasons |
1. Tuttnauer EHS is not drying properly 1) Check water level in the vacuum pump water reservior, no water will cause the dry stoping 2) Check the vacuum pump valve (15) to allow water to the pump 3) Check the vacuum valve (52) is open allowing the pump to draw from the chamber 4) Check the load Maximum Solid Load: 2540 EHS: 13.2 Lb 3870 EHS: 44.1 Lb Maximum Porous load: 2540 EHS: 4.4 Lb 3870 EHS: 15.5 Lb |
|
|||||
The technical information is just to help sterilizer users as reference to do their trouble shooting. no guarranty implied. For each specific technical help to do with sterilizers or autoclaves, go to www.sterilizers.com ... or call 1-800-801-9934 Contact Us |