Tuttnauer EHS "High Press" is displayed on the screen |
Problem Description:
Tuttnauer EHS "High Press" is displayed on the screen |
Reasons |
1. Tuttnauer EHS "High Press" is displayed on the screen in sterilizing stage The by-pass needle valve besides the steam trap is clogged. 1) Clean it 2) Close it completely then make a half a turn |
|
|||||