Tuttnauer EHS fast exhaust valve (73) is always open or closed
Problem Description:
Tuttnauer EHS fast exhaust valve (73) is always open or closed
Reasons
1. Tuttnauer EHS fast exhaust valve (73) is always open or closed

1) Check the JP10-12 on the ANL-T2 board
2) Check the coil of the solenoid
3) Check the mechanism of the solenoid to make sure no dirty inside
4) If the above does not work, then think about the digital board DIG-T2