Sterilizer Manufacturer: Pelton & Crane |
Model of Sterilizer: OCR |
Problem Title:
Heating on OCR autoclave |
Problem Description:
I am working on this unit and it is getting heat but after 1 1/2 hours the temp is only at 240 and the psi is about 8. I don't know if the heating elements need to be replaced or if I need to look at something else. If it is the heating elements what is the part # and does it include all 3 elements or 1 each. |
Reasons |
1. Is this an original machine or does it have an autovent function? |
Solutions |
1. You can test the heater elements ohm readings. They should read about 11-12 ohms |
|
|||||