Sterilizer Manufacturer: Scican Statim |
Model of Sterilizer: Statim 2000 |
Problem Title:
code 6 : cycle aborted, " not sterile" three times |
Problem Description:
code 6 : cycle aborted, " not sterile" three times |
Reasons |
1. The software has deteced a steam generator temperature 5 degrees centigrade higher then the chamber within 7,2 seconds after a purge during the sterilization cycle. |
Solutions |
1. Calibrate boiler. If problem persists, replace and calibrate boiler. |
|
|||||