Problem Title:
code 6 : cycle aborted, " not sterile" three times
Problem Description:
code 6 : cycle aborted, " not sterile" three times
Reasons
1. The software has deteced a steam generator temperature 5 degrees centigrade higher then the chamber within 7,2 seconds after a purge during the sterilization cycle.

Solutions

1. Calibrate boiler. If problem persists, replace and calibrate boiler.