Sterilizer Manufacturer: Scican Statim |
Model of Sterilizer: |
Problem Title:
Error 25 |
Problem Description:
Statim 5000 showing Error 25 |
Reasons |
1. The software has failed to detect a need to pump water in 90 seconds. |
Solutions |
1. Check boiler thermal fuse. 1 ohm reading or less is good. If thermal fuse reads good,check for appox. 11 ohms reading between terminals J1-3 & J1-4. Also check for line voltage at the same terminals. If these test show as good, then the problem is with the PCB. |
|
|||||