Sterilizer.com Sterilizer Parts Technical Home FAQ Search Submit a Question Your Local Service Login
Enter question keywords
to search

Sterilizer Manufacturer:
Scican Statim
Model of Sterilizer:
Problem Title:
Error 25
Problem Description:
Statim 5000 showing Error 25
Reasons
1. The software has failed to detect a need to pump water in 90 seconds.

Solutions

1. Check boiler thermal fuse. 1 ohm reading or less is good. If thermal fuse reads good,check for appox. 11 ohms reading between terminals J1-3 & J1-4. Also check for line voltage at the same terminals. If these test show as good, then the problem is with the PCB.

The technical information is just to help sterilizer users as reference to do their trouble shooting.
no guarranty implied. For each specific technical help to do with sterilizers
or autoclaves, go to www.sterilizers.com ... or call 1-800-801-9934 Contact Us