Sterilizer Manufacturer: Tuttnauer |
Model of Sterilizer: |
Problem Title:
Tuttnauer E Water Pump machine default time settings |
Problem Description:
Tuttnauer E Water Pump machine default time settings |
Reasons |
1. Tuttnauer E Water Pump machine default time settings 2340 = 30 Sec 2540 = 35 Sec 3870 = 65 Sec The above datum just for reference only. After the machine runs for a long time, you may need to increase the time more to make sure enough water filled in the chamber To reset the time, just use the following step: 1) Press and STOP repeatedly until you see Code: XXX 2) use UP/DN to get 105, the press STOP again 3) you get message Water in = XXX sec 4) using the UP/DN to set new times of seconds 5) The Ea Type shows up. using UP/DN to set 0 for E/EK; and 1 for EA/EKA (with Air pump for drying) 6) press STOP again |
|
|||||
The technical information is just to help sterilizer users as reference to do their trouble shooting. no guarranty implied. For each specific technical help to do with sterilizers or autoclaves, go to www.sterilizers.com ... or call 1-800-801-9934 Contact Us |