Tuttnauer E Water Pump machine default time settings
Problem Description:
Tuttnauer E Water Pump machine default time settings
Reasons
1. Tuttnauer E Water Pump machine default time settings

2340 = 30 Sec
2540 = 35 Sec
3870 = 65 Sec

The above datum just for reference only. After the machine runs for a long time, you may need to increase the time more to make sure enough water filled in the chamber To reset the time, just use the following step:

1) Press and STOP repeatedly until you see Code: XXX
2) use UP/DN to get 105, the press STOP again
3) you get message Water in = XXX sec
4) using the UP/DN to set new times of seconds
5) The Ea Type shows up. using UP/DN to set 0 for E/EK; and 1 for EA/EKA (with Air pump for drying)
6) press STOP again

The technical information is just to help sterilizer users as reference to do their trouble shooting.